Disposable ETT cuff manometer

DrParasite

The fire extinguisher is not just for show
6,196
2,052
113
Our company just gets the best deal they kind find from our supplier that the crews don’t complain about.
apples and oranges. if your company is providing the equipment, than your agency has approved it's use.
The majority of people use their own stethoscope and more people are starting to carry the raptor shears.
So did mine... our policies clearly said that each person was permitted to use their own scope.
Heck, I’m one of the few racetrack certified medics we have and our medical director doesn’t know how to use some of the special equipment that we carry. The supervisors have literally said “if there is any piece of equipment that will help you do your job at the track just let me know and we will get it”.
if your supervisors/management/company/agency are providing that equipment, than they are permitting you to use it. that's much different than bringing your own stuff from home and using it on the job.
I know there are many cops and firefighters who carry small pieces of personal gear that they prefer to use. I’m not talking about bringing in my own LP15 to work.
I know many EMS people that would carry expandable batons, handcuffs, flashlights, and couple that would carry their personal firearms, if their agency would allow them to. carrying your own medical equipment is much different than carrying a personal tool to make your job a little easier.
Classic EMS doomsday lines. Do you have any proof of this ever happening? Any examples? Maybe some relevant case law?
sure. I called up my agency's insurance agent (who, by his very nature, is a subject matter expert on insurance, and on what insurance policies cover), and asked if they would back me if I used unapproved equipment during my job. They flatout told me no, because i would be operating outside the conditions set forth by my employer, which is what the insurance policy applies to. They will cover me if I am doing my job, as specified by my employer, if I am following the guidelines set forth by my employer (which includes using the appropriate equipment, set forth by my employer). If I am operating outside of those guidelines, than any liability would be my responsibility only, not the agency.

Why don't you do the same? Get the company name, and or the agent, and ask if you can obtain non-company provided or approved equipment to use it on your patients. Let me know what he says.
I'm sure not using an agency supplied stethoscope opens me up to incredible liability. I mean it's not like my agency approved it, how do they know I can even hear lung sounds??
The appropriate response is, "it depends." does your agency permit you to use a personal stethoscope? meaning, do their written equipment/uniform requirements permit the use of personal scopes? Most do.

If you didn't like the brand of IV catheters that your agency provided, could you order your own and use them on your patients? What about NPAs? If your agency uses combitubes as your BIAD, but you think they suck, would they have any objections to you ordering some Igel from the local medical supply company?
 

Carlos Danger

Forum Deputy Chief
Premium Member
4,510
3,234
113
If you didn't like the brand of IV catheters that your agency provided, could you order your own and use them on your patients? What about NPAs? If your agency uses combitubes as your BIAD, but you think they suck, would they have any objections to you ordering some Igel from the local medical supply company?

Well, there's a big difference between invasive medical supplies that most individuals can't even legally purchase never mind use on their own, like an IV catheter or an LMA, and something like a stethoscope than anyone can buy in any corner drug store.

Not to mention the fact that liability depends first and foremost on the action actually causing damages. How could using one's personal stethoscope instead of the ones provided by your agency ever be shown to be the proximate cause of a bad outcome? Especially when personal ones are typically higher quality than the cheapos that most EMS agencies provide. How much effort can you foresee a plaintiff's attorney devoting making that case?

Where does this reasoning end? Do individuals need to fear liability for using personal watches on duty to time drug administrations? Calculators on their phones to figure dosages?

Can you provide an actual example of a time that an agency or one of their employees was found liable for damages incurred because an employee used a personal stethoscope or cuff manometer that wasn't authorized by policy?
 

DrParasite

The fire extinguisher is not just for show
6,196
2,052
113
Well, there's a big difference between invasive medical supplies that most individuals can't even legally purchase never mind use on their own, like an IV catheter or an LMA, and something like a stethoscope than anyone can buy in any corner drug store.
your right; I'm not the one who said anything about stethoscopes.
Can you provide an actual example of a time that an agency or one of their employees was found liable for damages incurred because an employee used a personal stethoscope or cuff manometer that wasn't authorized by policy?
i've never heard of anyone ever using a personal cuff manometer.

and I would imagine most agencies permit personal stethoscopes, in writing, in their written procedures.
 

DesertMedic66

Forum Troll
11,268
3,450
113
your right; I'm not the one who said anything about stethoscopes.i've never heard of anyone ever using a personal cuff manometer.

and I would imagine most agencies permit personal stethoscopes, in writing, in their written procedures.
To my knowledge my agency has zero protocol/policy about personal stethoscopes. The only thing that is required is for 1-2 stethoscopes be on the ambulance. It is really a not needed policy.
 

Tigger

Dodges Pucks
Community Leader
7,844
2,794
113
apples and oranges. if your company is providing the equipment, than your agency has approved it's use.So did mine... our policies clearly said that each person was permitted to use their own scope.if your supervisors/management/company/agency are providing that equipment, than they are permitting you to use it. that's much different than bringing your own stuff from home and using it on the job.I know many EMS people that would carry expandable batons, handcuffs, flashlights, and couple that would carry their personal firearms, if their agency would allow them to. carrying your own medical equipment is much different than carrying a personal tool to make your job a little easier.sure. I called up my agency's insurance agent (who, by his very nature, is a subject matter expert on insurance, and on what insurance policies cover), and asked if they would back me if I used unapproved equipment during my job. They flatout told me no, because i would be operating outside the conditions set forth by my employer, which is what the insurance policy applies to. They will cover me if I am doing my job, as specified by my employer, if I am following the guidelines set forth by my employer (which includes using the appropriate equipment, set forth by my employer). If I am operating outside of those guidelines, than any liability would be my responsibility only, not the agency.

Why don't you do the same? Get the company name, and or the agent, and ask if you can obtain non-company provided or approved equipment to use it on your patients. Let me know what he says.The appropriate response is, "it depends." does your agency permit you to use a personal stethoscope? meaning, do their written equipment/uniform requirements permit the use of personal scopes? Most do.

If you didn't like the brand of IV catheters that your agency provided, could you order your own and use them on your patients? What about NPAs? If your agency uses combitubes as your BIAD, but you think they suck, would they have any objections to you ordering some Igel from the local medical supply company?
Perhaps some research on logical fallacies would help, here's a primer. https://owl.english.purdue.edu/owl/resource/659/03/

Until then, I'll wait for some case law.
 

ISRDEU

Forum Ride Along
4
0
1
Michelle Eibell

USA - Mercury Medical

11300 49th Street North,

Clearwater, Florida USA 33762-4807

Toll Free: (800) 237-6418 (USA only)

Tel: (727) 573-0088

Fax: (727) 571-3922
 

CWATT

Forum Lieutenant
182
50
28
@Remi would probably be able to answer this, but is there really a big difference between using a specific cuff pressure compared to slowly inflating the cuff until you no longer have air leak and stopping right there? I have never personally seen a manometer used to measure cuff pressures in the ED, PICU, ICUs, or ORs.

We check it on every intubated patient.

The modern high volume / low pressure ETT cuffs are designed such that if the tube is properly sized and the cuff is not overfilled, cuff pressures above about 20 are unlikely.

I picked up a patient last week whose cuff pressure was about 60cmH20.
 

Brandon O

Puzzled by facies
1,718
337
83
@Remi would probably be able to answer this, but is there really a big difference between using a specific cuff pressure compared to slowly inflating the cuff until you no longer have air leak and stopping right there? I have never personally seen a manometer used to measure cuff pressures in the ED, PICU, ICUs, or ORs.

I'd say it is more or less the standard of care in the ICU.
 

Peak

ED/Prehospital Registered Nurse
1,023
604
113
I'd say it is more or less the standard of care in the ICU.

We just deflate the cuff until you hear air leak or have increased leak on the vent, typically adding about 0.5-1 mL of air, and call it good. Our view is that if you have to use more air than the standard cuff volume then your tube isn't correctly sized or they is a different pathology (ARDS, et cetera) that is causing us to need so much pressure in which case we need to address that by changing our vent mode/settings, changing our medications, change other treatments like proning the patient, or considering ECMO as a gas exchange adjunct. We haven't had any problems with impaired tracheal perfusion (well, outside of the NICU but that is of course not relevant to cuff pressures) so I think we haven't really had a need to address cuff pressures in a more precise way. Granted I have only worked in hospital systems that have ECMO, jet vents, oscillators, et cetera and in the field (I never ran IFTs) cuff pressure typically wasn't high on my list of concerns if I tubed somebody.
 

Brandon O

Puzzled by facies
1,718
337
83
We just deflate the cuff until you hear air leak or have increased leak on the vent, typically adding about 0.5-1 mL of air, and call it good. Our view is that if you have to use more air than the standard cuff volume then your tube isn't correctly sized

It's not really a matter of size. We don't select tube size to match trachea size, like fitting a hand into a glove, other than in a very general sense ("Can't seem to cram an 8.0 into this one, let's try smaller"). This is because the cuff molds to fit the surrounding airway. Since that may require a larger or smaller volume, cuff inflation volume is really not predictable. That's why pressure is a better endpoint.

I suppose you can argue that if you inflate to the correct pressure and are still seeing cuff leaks, then you need to upsize -- but as cuffs can usually be inflated to obscenely large volumes if you so desire, I'm not sure how true that is.

If your airway pressures are consistently exceeding the pressure in the cuff, then physics dictates that you will probably have some leak. Expunging this by inflating the cuff to unsafe pressures would not be defensible. Mucosal ischemia is not a common thing in this era, but that is largely because of these basic good practices, and for that very reason I think it would be hard to defend your system if you weren't following them and had a bad outcome.

Granted I have only worked in hospital systems that have ECMO, jet vents, oscillators, et cetera and in the field (I never ran IFTs) cuff pressure typically wasn't high on my list of concerns if I tubed somebody.

It's not really a factor of vent strategy. Safe airway management stands on its own.

I agree that in the field this stuff is probably not generally necessary, and using a fingerometer on the balloon is probably reasonable.
 

Peak

ED/Prehospital Registered Nurse
1,023
604
113
It's not really a matter of size. We don't select tube size to match trachea size, like fitting a hand into a glove, other than in a very general sense ("Can't seem to cram an 8.0 into this one, let's try smaller").

I'm going to disagree with you on this. A tube that is too small will either need a higher ramp, more pressure, or increased duration of inspiratory/expiratory phase compared to a larger tube. This is certainly much more pronounced for say a 2.0mm vs 2.5mm than 6.5mm vs 8.0mm, but a 6'5" guy should certainly be getting a bigger tube than a 55 kg grandma.

is because the cuff molds to fit the surrounding airway. Since that may require a larger or smaller volume, cuff inflation volume is really not predictable. That's why pressure is a better endpoint.

Just for the sake of clarity I want to reinforce that I do not endorse fully inflating the cuff with the manufacturer's printed volume. I prefer to inflate the cuff slowly while bagging the patient and to stop adding more volume to the cuff as soon as there is no audible evidence of air leak. Often this is a small fraction of the volume that the manufacturer has printed on the tube/packaging.

If your airway pressures are consistently exceeding the pressure in the cuff, then physics dictates that you will probably have some leak. Expunging this by inflating the cuff to unsafe pressures would not be defensible. Mucosal ischemia is not a common thing in this era, but that is largely because of these basic good practices, and for that very reason I think it would be hard to defend your system if you weren't following them and had a bad outcome.

If your airway pressures are exceeding the pressures in the cuff, and your airways pressures are exceedingly high, then the cause of that increased airway pressure needs to be addressed. I would never simply further inflate a cuff because their airway pressures are higher, any good clinician should be assessing and treating the underlying cause (mucous plugs, pathological lung disease, inappropriate volumes, kinked tubing, et cetera). I don't think that inflating/deflating cuffs based on a physical assessment increases liability over using a manometer.

For the sake of argument am I at higher liability if I place an art line without ultrasound guidance if I could identify the anatomy without one? Am I at higher liability if I intubate a neo with direct laryngoscopy instead of using a glidescope? I don't think adding tools necessarily decreases risk or liability.

It's not really a factor of vent strategy. Safe airway management stands on its own.

If your patient has an ineffective ventilation/respiration management then it needs to be addressed. There is much more that goes into gas exchange than simply the size of a tube, how a cuff is inflated, or how a tube is secured. Different pathological or physiological processes don't have a one size fits all vent setting, and sometimes that means changing your management so that you can better facilitate gas exchange.

I agree that in the field this stuff is probably not generally necessary, and using a fingerometer on the balloon is probably reasonable.

I don't advocate for trying to figure out the pressure in the tube by pressing on the balloon. I can understand the potential desire for IFT crews for a manomoter, especially with the potential of scene/transport noise making a physical assessment more difficult, but I don't think that 911 only groups need one. When I was in fire I got one cabinet for medical equipment (on the engine anyway) and I can think of all kinds of stuff I would prefer to have, now that I'm doing more wilderness-ish stuff where we carry in everything ourselves ounces add up to pounds real quick. I do think that the gold standard should always be a good physical assessment, and given my experience in the medical field so far I don't think it would have been a benefit in my practice.
 
Last edited:

Brandon O

Puzzled by facies
1,718
337
83
I'm going to disagree with you on this. A tube that is too small will either need a higher ramp, more pressure, or increased duration of inspiratory/expiratory phase compared to a larger tube. This is certainly much more pronounced for say a 2.0mm vs 2.5mm than 6.5mm vs 8.0mm, but a 6'5" guy should certainly be getting a bigger tube than a 55 kg grandma.

Yes, a tighter tube make life a little more annoying, requiring higher peaks, making it difficult to bronch, etc. The answer is to place the largest tube you can fit without needing one of those post-diggers. (Most folks can take an 8.0.) I suppose you could view this as matching the size to the patient, but in general it can be done empirically, at least in adults.

Just for the sake of clarity I want to reinforce that I do not endorse fully inflating the cuff with the manufacturer's printed volume. I prefer to inflate the cuff slowly while bagging the patient and to stop adding more volume to the cuff as soon as there is no audible evidence of air leak. Often this is a small fraction of the volume that the manufacturer has printed on the tube/packaging.

Other than the approximate ballpark I would ignore the "recommended" volume as well (I have no idea what it says on the package for our tubes). I agree that cuff leak is an important endpoint to recognize, although we use the vent, which is more exact. But the other endpoint is cuff pressure.

I think the point is that you might conceivably have a cuff leak even in the presence of a hyperinflated cuff at unsafe pressure. This might be due to atypical airway anatomy or an imperfectly-positioned tube. While it should make sense that a "barely sealed" cuff-to-trachea interface should equate with low pressures, this should only be true if that interface is perfectly homogenous. Since that is not always true, there could be high and low points, equating to high pressures while still allowing a leak. (There was some talk of using "tapered" cuffs lately to try and match surfaces even better, which hasn't really panned out.)

If your airway pressures are exceeding the pressures in the cuff, and your airways pressures are exceedingly high, then the cause of that increased airway pressure needs to be addressed.

Poor compliance is certainly suggestive of lung pathology, but mostly as a marker; we don't manage pressures (other than trying to achieve safe plateaus), we manage disease states.

For the sake of argument am I at higher liability if I place an art line without ultrasound guidance if I could identify the anatomy without one? Am I at higher liability if I intubate a neo with direct laryngoscopy instead of using a glidescope? I don't think adding tools necessarily decreases risk or liability.

I don't think so yet, but we may be getting there. The missing link is data suggesting clear superiority in patient outcomes. There IS good data that ultrasound increases success and decreases complications in line placement, but not really that it affects many patient-centered outcomes like mortality, so you can still get away with blind placement. There is NOT clear data yet that VL is clearly superior to DL.

A good analogy might be central line placement in the IJ. While there is not, to my knowledge, a head-to-head study on this, everyone has pretty much abandoned doing these by the landmark approach, because they are usually extremely easy under ultrasound, whereas accidentally puncturing a carotid is very undesirable. That combination makes it pretty undefensible if you wreck a carotid when an ultrasound was waiting in the closet 20 feet away. Similarly, in resource-rich settings when the equipment and personnel are readily available, I think there is little to be argued against simple cuff pressure checks to prevent the really gnarly outcomes like tracheoesophageal fistulas.

If your patient has an ineffective ventilation/respiration management then it needs to be addressed. There is much more that goes into gas exchange than simply the size of a tube, how a cuff is inflated, or how a tube is secured.

I would say these have next to nothing to do with gas exchange.
 

MSDeltaFlt

RRT/NRP
1,422
35
48
Me, personally, minimal leak is good for prehospital. When I have the ICU as an RT I like to use cuff pressure.

As far as ETT's go, since this particular thread has evolved, I choose my ETT size based on 0.5 smaller than the largest tube that will fit. That way you get the lowest intrapulmonary pressures possible and can use just enough air to inflate the cuff to get rid of any wrinkles on the bulb to prevent any hard edges on the inner lining of the trachea. What that can also do on weaning (I know this is an EMS website and not an RT website and we do not wean in the field, but...) is to allow the patient to breathe around the tube when you're doing a leak test to assess spontaneous respirations just prior to extubation.

My personal anecdotal opinion is if you need to use a syringe larger that a 3cc or 5cc to inflate the cuff then your tube is too small. Remember, tracheas are larger in diameter compared to ETT's than you might think.
 

Brandon O

Puzzled by facies
1,718
337
83
As far as ETT's go, since this particular thread has evolved, I choose my ETT size based on 0.5 smaller than the largest tube that will fit.

I assume you mean this more as a general concept (aiming to place a tube a little smaller than one you think you'd have to cram in) than a practical one -- I assume you are not intubating by passing larger and larger tubes until you reach one that doesn't fit, then downsizing by one...
 

MSDeltaFlt

RRT/NRP
1,422
35
48
I assume you mean this more as a general concept (aiming to place a tube a little smaller than one you think you'd have to cram in) than a practical one -- I assume you are not intubating by passing larger and larger tubes until you reach one that doesn't fit, then downsizing by one...


Uh, negative, Ghostrider. I'm referring to first pass scenarios. As I mentioned before, tracheas are larger than most might realize. Take me for example. I am a 48yo male who is 5'10 1/2'' - 5'11" depending on how angry you make me. And the inner diameter of my trachea is larger than the outer diameter of a 10.0 ETT. People need to treat tracheas in the same way they treat veins. You go larger for a reason. Advocate for your patient and give them the most flow with the lowest pressures possible.

Drop it like it's hot.
 

ThadeusJ

Forum Lieutenant
240
69
28
You go larger for a reason. Advocate for your patient and give them the most flow with the lowest pressures possible.

The maximum airflow that can pass through a tube is an inverse function of the radius to the power of four. In English, that means that if you reduce the radius of a tube by half, the flow that can pass through the tube is cut 16 times (1/16). For patients that have respiratory issues to begin with (e.g. COPD), trying to get them to breathe spontaneously through the tube prior to extubation can be a non-starter as the resistance it way too high. I don't know how many times I had to explain to a doctor that if they could breathe through the pea shooter they put in, the patient wouldn't be in the hospital to begin with.
 

VFlutter

Flight Nurse
3,728
1,264
113
That's what SBT with Pressure Support is for....
 
Top