The maximum airflow that can pass through a tube is an inverse function of the radius to the power of four. In English, that means that if you reduce the radius of a tube by half, the flow that can pass through the tube is cut 16 times (1/16). For patients that have respiratory issues to begin with (e.g. COPD), trying to get them to breathe spontaneously through the tube prior to extubation can be a non-starter as the resistance it way too high. I don't know how many times I had to explain to a doctor that if they could breathe through the pea shooter they put in, the patient wouldn't be in the hospital to begin with.
That's what SBT with Pressure Support is for....
Yes, with the advancement of modern microprocessor ventilators, T-pieces are no longer needed prior to extubation. However, with the smaller tubes the PIP's are unnecessarily high and give false readings as to what is going on in the pulmonary tissue. The lower the pressures + the easier the flow = the less resistance = the less chance for barotrauma.
You choose you ETT size just like you choose your IV cath size. Get as big as you can get for very specific reasons.
It is never a "go big or go home" philosophy. It's more like a "if it'll fit it stick it" but just one size smaller that the absolute largest that will fit. After SBT's they sometimes like to deflate the cuff to make sure move enough air around the tube just prior to extubation. If after a few breaths they're still doing good, have them cough as you pull it. But while they're intubated bigger is better.