How accurate are the LifePak/Zoll Defibs at Identifying Rhythms?

MMiz

I put the M in EMTLife
Community Leader
5,551
412
83
This thread got me thinking. Based on your experiences, how accurate are current defibrillators at identifying rhythms?

Would you trust a BLS provider to treat a patient based on the defibrillators interpretation alone?
 

silver

Forum Asst. Chief
916
125
43
"A total of 3,448 AED rhythms were available for interpretation. Sensitivity and specificity for appropriate AED management of a shockable (VF or VT) rhythm were 81.0% (95% CI 77.9% to 83.8%) and 99.9% (95% CI 99.7% to 100%), respectively. Positive and negative predictive values were 99.6% (95% CI 98.7% to 99.9%) and 95.5% (95% CI 94.7% to 96.2%), respectively. There were 132 errors associated with AED management. Two errors resulted in delivery of an inappropriate shock. In the remaining 130 errors, a shockable rhythm was not shocked. Fifty-five (42.3%) errors were AED dependent, 70 (53.9%) were operator dependent, and 5 (3.9%) were unclassified."

http://www.ncbi.nlm.nih.gov/pubmed/11524645

Potential limitations of this study: technology used is currently 10+ years old and that is with AEDs only. Obviously the manual defibs used today have better algorithms.


Edit: Oh, another experience/opinion thread. I'm just going to take my science elsewhere... :sad:
 
Last edited by a moderator:

bstone

Forum Deputy Chief
2,066
1
0
"A total of 3,448 AED rhythms were available for interpretation. Sensitivity and specificity for appropriate AED management of a shockable (VF or VT) rhythm were 81.0% (95% CI 77.9% to 83.8%) and 99.9% (95% CI 99.7% to 100%), respectively. Positive and negative predictive values were 99.6% (95% CI 98.7% to 99.9%) and 95.5% (95% CI 94.7% to 96.2%), respectively. There were 132 errors associated with AED management. Two errors resulted in delivery of an inappropriate shock. In the remaining 130 errors, a shockable rhythm was not shocked. Fifty-five (42.3%) errors were AED dependent, 70 (53.9%) were operator dependent, and 5 (3.9%) were unclassified."

http://www.ncbi.nlm.nih.gov/pubmed/11524645

Potential limitations of this study: technology used is currently 10+ years old and that is with AEDs only. Obviously the manual defibs used today have better algorithms.


Edit: Oh, another experience/opinion thread. I'm just going to take my science elsewhere... :sad:

From Ann Emerg Med. 2001. Anything newer? Hopefully things have improved in the 10+ years since this study came out.
 
OP
OP
MMiz

MMiz

I put the M in EMTLife
Community Leader
5,551
412
83
I'm thinking not about the AED function, but when using the monitor as a monitor during a cardiac code. I remember that there was a code review/interpretation feature. I'm wondering if they were accurate.
 

EMT PR

Forum Ride Along
7
0
0
don't count on those...

According to my experience (with LP and PROTOCOL) don't trust on those… some time they wrong- but most of the time they are pretty accurate.
You can use them- but don't relate on them as your only diagnosis
 

Handsome Robb

Youngin'
Premium Member
9,736
1,174
113
Are you talking about basic rhythm interp or 12-lead interp?

If you are letting the monitor tell you what's going on in a 12-lead you have serious issues.
 

EMT PR

Forum Ride Along
7
0
0
both

both- but the self interpretation of 12 lead is more offten incorrect then the monitor interpretation.
 

TomB

Forum Captain
393
82
28
The GE-Marquette 12SL interpretive algorithm does a fairly decent job identifying conduction abnormalities like bundle branch blocks and has a fairly high specificity when it gives the ***ACUTE MI SUSPECTED*** or ***ACUTE MI*** message. However, it does not do particularly well at basic rhythm analysis. One of the reasons the sensitivity for acute STEMI is somewhat low is that the STEMI criterion (1 mm of ST-elevation in 2 or more anatomically continguous leads -- 2 mm for leads V2 and V3) is problematic because it does not take into account the rule of proportionality. In other words, it does not consider the depth of the S-wave. If you're programming a computer it's easy to make it follow rules. It's hard to program it to sometimes ignore the rules. It's also important to point out that computerized interpretive algorithms are very susceptible to errors from poor data quality.
 

Flight-LP

Forum Deputy Chief
1,548
16
38
I'm thinking not about the AED function, but when using the monitor as a monitor during a cardiac code. I remember that there was a code review/interpretation feature. I'm wondering if they were accurate.

My question, again, is why is it necessary?

The EMT-B has no business even attempting to interpret a rhythm. If the pulse is missing, hook them up to the computer who will interpret if a shock is needed or not. It really is that simple. The AED was not designed to be used in an ALS environment. Now Zoll has created an "all in one unit" in their "M" series, that is fine. The unit was designed to allow integration at a reduced cost so one unit could be used by different level personnel. It was not designed to allow a non-trained BLS provider to have a screen for an interpretation that they are not trained or qualified to perform.

The old LP-300 was a prime example. It had a manual override feature that was misused and gave a false sense of BLS autonomy.

This is an excellent example of the age old issue in EMS, a BLS provider does not need any additional tools that they are not prepared, nor trained to use. Remember the KISS principle, if we as an industry actually followed it, we probably wouldn't be worried about such trivial issues.

Now on the manual monitor / defibrillator side, personally I could care less what the interpretation says on the strip in the 12 lead or code review mode. I do not function like LaCo, I for one will actually print a strip, measure out my waves and complexes, and do a manual interpretation. It is my sincerest hope that other medics do as well.

I like how Tom put it, its easy for the computer to follow the rules, but hard to deviate. For the layperson and BLS level, this is a fine way to deliver resuscitative care. At the ALS level though, a little critical thinking and creativity needs to come into play, hence why we need to move away from the computer.

So Matt, yes I would trust the computer for a BLS provider, a lot more than I would trust a manual interpretation.
 

Flight-LP

Forum Deputy Chief
1,548
16
38
both- but the self interpretation of 12 lead is more offten incorrect then the monitor interpretation.

Not if you have a medic who actually knows what they are doing. If a self interpretation is "more offten incorrect", then your QA and clinical department has some work ahead.....
 

mikie

Forum Lurker
1,071
1
36
on a side note...

The GE-Marquette 12SL interpretive algorithm...

Is that what is standard across all the models? LP, Zoll, Phillips?

Can you point me in the direction of finding more information on the actual 'algorithm' for how the EKG interpretation does occur throughout the various machines (LifePak mostly)? Google-ing wasn't so helpful..

Thanks
 

TomB

Forum Captain
393
82
28
Yes for LP12 and ZOLL M-series. No for Philips MRx. Not sure about LP15.

Tom

Is that what is standard across all the models? LP, Zoll, Phillips?

Can you point me in the direction of finding more information on the actual 'algorithm' for how the EKG interpretation does occur throughout the various machines (LifePak mostly)? Google-ing wasn't so helpful..

Thanks
 

TheGodfather

Forum Lieutenant
165
0
0
I for one will actually print a strip, measure out my waves and complexes, and do a manual interpretation. It is my sincerest hope that other medics do as well.

I completely agree with this statement. I keep a pocket ECG ruler in the back pouch of my monitor. :beerchug:
 

EMSrush

Forum Captain
349
0
0
I was just wondering about this myself not too long ago... I was wondering if the LP-12 is able to differentiate between an acute STEMI and an imposter? Will it interpret all elevation as a STEMI even if the elevation is benign?
 

TheGodfather

Forum Lieutenant
165
0
0
LP-12 has a 98% specifity... IE, if it says **acute mi suspected** , 98% of the time it is right... however, along with that, it has approx a 52% sensitivity, IE; 48% of the time, it overlooks the problem...

moral, if it says **acute mi suspected** STONGLY consider it... however, do not count on it finding it approx 1/2 of the time... measure, measure, measure!
 

Brandon O

Puzzled by facies
1,718
337
83
Zoll is switching to an in-house algorithm for their new Propaqs. The M and E will still use the Marquette.

Marquette is surprisingly good at sussing out STEMI mimics but not perfect. Specificity numbers vary by study but are 90%-100% with good data. However pretest probability affects this significantly. (Can't get into Bayesianism on my phone keyboard!) In sum, good tool, needs overall clinical and electrical context like all tools. We shouldn't be reliant on nor ignore it but understand and use it appropriately.

via Droid
 

tacitblue

Forum Crew Member
65
0
0
This thread got me thinking. Based on your experiences, how accurate are current defibrillators at identifying rhythms?

Would you trust a BLS provider to treat a patient based on the defibrillators interpretation alone?

I use Phillips MRx at my company, and I never trust its interpretation of the 3 or 12 lead. Why? The VF alarm comes on when my patients move. The extreme brady alarm goes off with paced patients with good mechanical capture and normal perfusion. When interpreting rhythms, I almost always print a strip, calculate the rate and measure the intervals myself. On 12 leads I interpret axis and R wave progression, measure ST deviations, and look for BBBs myself as well. These are good habits and every medic worth his salt should be very good at this.

As for EMTs on a cardiac arrest deciding to defibrillate or not, I would actually rather the EMT make the decision on whether it was VF/VT vs asystole or PEA him/herself. It isn't that hard and if trained, they could probably do better than the machine. This isn't in their scope of practice in my area nor do they have the education to do so, so its moot. As an aside, I have told my EMT partner that if I am not around and we are holding the wall at the ED or what not, she is to use the AED function on the monitor.
 
Last edited by a moderator:

rmabrey

Forum Asst. Chief
854
2
18
both- but the self interpretation of 12 lead is more offten incorrect then the monitor interpretation.

A cardiologists at a local specialty hospital called us to pick up a patient having an MI cause the monitor interpreted MI. Needless to say nobody else saw MI when interpreting the 12-lead, and our LP15's didn't even kick MI.

Sent from my Desire HD using Tapatalk
 
Top