I'm used to Zoll but here at work we have Lifepak 12's.
I was out playing with it and noticed when I went to change the energy setting, it said Biphasic. However...it allowed me to turn the energy up to 360j.
Why? What? Huh?
I have the manual but I can't really find anything about this
Physio-Control recently joined the EMTLife Forum as a member of the new Verified Vendor Program and came across this discussion. We wanted to offer some additional information that should help provide context.
When treating difficult-to-defibrillate patients, escalation of energy can be an important element of improved outcomes. Physio-Control products have been studied in more than twice as many cardiac arrest patients as all other manufacturers combined. The data shows that the only way to increase shock efficacy when lower-energy shocks fail is to increase energy. That’s why we offer the highest available energy—up to 360J.
Many factors contribute to successful defibrillation, including: 1.) the peak current of the waveform, 2.) how long the current is applied, and 3.) how the current is maintained throughout the shock. While peak current plays a role, it does not determine shock efficacy. Energy includes all three elements and has been shown to best describe the therapeutic dose delivered to the heart.
Consider that three clinical studies have compared biphasic waveforms used by Physio-Control (biphasic truncated) and Zoll (biphasic rectilinear) in synchronized cardioversion. The combined results show that, though Zoll’s waveform delivers higher levels of current, the waveforms are equally effective at 200J (96% efficacy).
Some concern was expressed earlier in the thread with regards to the use of 360J. Bear in mind that both the 2010 American Heart Association and European Resuscitation Council Guidelines recommend energy protocols in which the initial dose follows manufacturers’ recommendations (120–200J) and subsequent shocks given at an equivalent or higher dose. For reference, “Human studies have not demonstrated evidence of harm from any biphasic waveform defibrillation energy up to 360 J…” (American Heart Association Guidelines 2010; pg. S708).
Additionally, independent defibrillation studies support the AHA Guidelines. Two clinical trials using escalating energy protocols to 360J reported no measures of myocardial damage by enzymes, ECG changes or ejection fractions (Stiell I, et al).
- A randomized comparison of fixed lower versus escalating higher energy levels for defibrillation in out-of-hospital cardiac arrest (Circulation, 2007; 115:1511-1517. Neal S, et al).
- Comparison of the efficacy and safety of two biphasic defibrillator waveforms for the conversion of atrial fibrillation to sinus rhythm (American Journal of Cardiology, 2003; 92:810-14).