To say
PEEP is CPAP with a rate, as MsDeltaFlight did
, is correct in that the term PEEP is associated with modes that provide a rate or assisted breaths for ventilation. CPAP is generally thought of as a mode by itself.
A lot of this depends on the machine design for delivery and the manufacturer's use of the terms PEEP and CPAP.
If it is a single limb ventilator, as many transport vents are, a resistive or PEEP valve may be used on exhalation to create the PEEP or CPAP from the forward flow on exhalation.
Even the Pulmonetics LTV transport ventilator used an external PEEP valve (resistive) until their 1200 series. This is common with many single limb transport ventilators. So essentially they are using a similar concept as the resistive valve and mask which is what prehospital is using.
However;
Although it is the most ubiquitous form of ventilatory support, positive end expiratory pressure (PEEP) remains a large area of confusion for most physicians. For a start, the term PEEP is an anachronism, as the positive pressure is actually applied throughout the respiratory cycle and is more correctly termed “continuous positive airway pressure (CPAP)”;
http://www.ccmtutorials.com/rs/PEEP/page1.htm
Good statement but it pertains to ICU or double limb circuit ventilators.
- Auto-PEEP is gas trapped in alveoli at end expiration, due to inadequate time for expiration, bronchoconstriction or mucus plugging. It increased the work of breathing.
Auto-PEEP is caused by gas trapped in alveoli at end expiration. This gas is not in equilibrium with the atmosphere and it exerts a positive pressure, increasing the work of breathing,
http://www.ccmtutorials.com/rs/PEEP/page7.htm
Mechanical ventilators or other external devices deliver from the top. Patients can do their own form of PEEP against the glottis by pursed lip breathing or babies use grunting.
Grunting results from the partial closure of the glottis during forced expiration in an effort to maintain FRC.
However, if it is the result of over ventilation mechanically, it is inadvertent Auto-PEEP and measurements are done religiously on some patients to monitor this.
Good case study on this concept:
[FONT=arial, helvetica]Cardiovascular Instability Caused by Inadvertent Positive End-Expiratory Pressure in a Patient with Panlobular Emphysema Receiving Mechanical Ventilation [/FONT]
http://www.ajronline.org/cgi/content/full/174/5/1339
Good brief overview and intro to mechanical ventilation:
http://www.ccmtutorials.com/rs/index.htm
Once they reach that point, they get an ETT. CPAP or even BiPAP is effective if it can be done early enough. Many of the prehospital and a few old ED machines can only reach an FiO2 of 0.30. Also, by design on these machines, the higher the FiO2, the less flow available to the patient due to the mechanisms of entrainment.
Asthmatics may also need HeliOx to assist in ventilation which can be done through a mask such as a NRBM or a few ventilators. At this time, CPAP/PEEP is provided through a ventilator designed to handle HeliOx.
BiPAP is also not the correct term unless you are using a Respironics machine. You can also get into many variations of two level delivery.
Some machines also use the terms EPAP instead of PEEP but may or may not use CPAP to designate the MODE of CPAP.
Good article about CPAP and PSV/PEEP.
Physiologic Effects of Noninvasive Ventilation during Acute Lung Injury
http://ajrccm.atsjournals.org/cgi/content/full/172/9/1112/
Effects of the Components of Positive Airway Pressure on Work of Breathing During Bronchospasm
http://www.medscape.com/viewarticle/470756_1
This is a good read for those working with "PEEP" valves (resistive devices) to deliver either PEEP or CPAP.
Barotrauma from CPAP Systems Lacking Pressure Relief
http://www.mdsr.ecri.org/summary/detail.aspx?doc_id=8145