So what are some causes and solutions to artifact and other bad EKG's?
Here is what I posted on another topic just to get us started.
Pre-hospital really suspect that they failed to properly clean/scuff the skin leading to a bunch of artifact. I see it way to often even with the non fire EMS. Wiping with an alcohol pad is not enough but is better than many who just place them over hair, dirt and all.
To do it right you need hair shaved, wipe the skin, then scuff it. Use something like 3m red dot skin prep. A dry 4x4 does not scuff enough.
http://www.medexsupply.com/products/...pRollwithD.htm
The next cause sadly is often the cables are not fully connected to the machine. I have gotten into an ambulance and medics had artifact and I just reached over and pushed cable connector further into its socket and amazingly got a clean 12/15 lead.
Another common cause is the cables developing cracks. A busy service really should replace the cables every few months, while a slow rural service might need to replace every 1-2 years.
And yet another problem especially with older machines is electrical interference. With some machines you really needed everything off in the ambulance.
Thats just a small list of causes.
Here is what I posted on another topic just to get us started.
Pre-hospital really suspect that they failed to properly clean/scuff the skin leading to a bunch of artifact. I see it way to often even with the non fire EMS. Wiping with an alcohol pad is not enough but is better than many who just place them over hair, dirt and all.
To do it right you need hair shaved, wipe the skin, then scuff it. Use something like 3m red dot skin prep. A dry 4x4 does not scuff enough.
http://www.medexsupply.com/products/...pRollwithD.htm
The next cause sadly is often the cables are not fully connected to the machine. I have gotten into an ambulance and medics had artifact and I just reached over and pushed cable connector further into its socket and amazingly got a clean 12/15 lead.
Another common cause is the cables developing cracks. A busy service really should replace the cables every few months, while a slow rural service might need to replace every 1-2 years.
And yet another problem especially with older machines is electrical interference. With some machines you really needed everything off in the ambulance.
Thats just a small list of causes.