Hey guys, couple questions about pacing with the LP12....
1) If you have only the pacing pads connected, but not the limb leads, you can still change the amps and pacing rate, correct? I was reading through the operator's manual, and I'm a little confused because it states, "If the monitor detects ECG leads off during pacing, pacing continues at a fixed rate until the ECG lead is reattached. During fixed rate pacing, the pacemaker delivers pulses at the set pace rate regardless of any intrinsic beats the patient may have." So, the way I understand this is that the LP12 can't read the patient's pulse rate w/o the limb leads attached, and will do non-demand pacing at whatever rate was set prior to the limb leads becoming detached.
2) If you pace with only the pacing pads but no limb leads connected, you won't see anything on the monitor, right?
Thanks!
1) If you have only the pacing pads connected, but not the limb leads, you can still change the amps and pacing rate, correct? I was reading through the operator's manual, and I'm a little confused because it states, "If the monitor detects ECG leads off during pacing, pacing continues at a fixed rate until the ECG lead is reattached. During fixed rate pacing, the pacemaker delivers pulses at the set pace rate regardless of any intrinsic beats the patient may have." So, the way I understand this is that the LP12 can't read the patient's pulse rate w/o the limb leads attached, and will do non-demand pacing at whatever rate was set prior to the limb leads becoming detached.
2) If you pace with only the pacing pads but no limb leads connected, you won't see anything on the monitor, right?